CN0273:低输入偏置电流和高交流共模抑制的高速FET输入仪表放大器
消耗积分:0 | 格式:pdf | 大小:207.93KB | 2021-03-23
分享资料个
This version (03 Jan 2021 21:58) was approved by Robin Getz.The Previously approved version (14 Aug 2017 08:30) is available.Diff
CN0273: High Speed FET Input Instrumentation Amplifier with Low Input Bias Current and High AC Common-Mode Rejection
OVERVIEW
The circuit is a high speed FET input, gain-of-5 instrumentation amplifier (in-amp) with a wide bandwidth (35 MHz) and excellent ac common-mode rejection, CMR, (55 dB at 10 MHz). The circuit is ideal for applications where a high input impedance, fast in-amp is required, including RF, video, optical signal sensing, and high speed instrumentation. The high CMR and bandwidth also makes it ideal as a wideband differential line receiver. Ideal as a wideband differential line receiver.
DEMO REQUIREMENTS
  • EVAL-CN0273-EB1Z circuit evaluation board
  • +/-5 V dual power supply
  • Signal generator with a balanced differential drive source
  • 3 SMA to BNC Connector
HARDWARE SETUP
Power Supply
 +/-5V supply should be supplied to the connector for proper operation of the evaluation board
Input Signal
 Signal Generator w/ Balanced Differential Drive Source can be used as an input source.
Peak to Peak Input Voltage Restriction
  • Maximum allowable differential input is ±3.5 V/G1, where G1 is the gain of the first stage.
There is a tradeoff between the maximum allowable differential input and the closed-loop gain of the first stage.
Common Mode Input Voltage Restriction
  • Allowable input common-mode range for the composite circuit is from −3.5 V to +2.2 V
Bandwidth Input Restriction
  • Allowable input frequency bandwidth is 35MHz @ -3dB, typical for G = 5
Output Measurement
 Oscilloscope can be used to measure the output of the evaluation board with a gain of 14db below 10Mhz input as indicited in the graph below.
The output swing of the circuit is limited to ±3.5 V when operating on ±5 V supplies
ADVANCED CONFIGURATION
FET Amplifier Input Gain Stage
  • The circuit has a differential inputs gain of 1 + 2RF/RG
    • Where,
      • RF = Resistor R5 and R6
      • RG = Resistor R7
  • For this circuit, a feedback resistor of 200 Ω was chosen. The feedback capacitor, CF, was 2 pF for the best bandwidth flatness.
  • When the inputs are common-mode, there is no current flowing through the RG gain resistor. Thus, the circuit acts as a buffer for the common-mode inputs.
SCHEMATICS, PCB LAYOUT, BILL OF MATERIALS
End of Document