EVAL-CN0398-ARDZ屏蔽概述
消耗积分:2 | 格式:pdf | 大小:330.79KB | 2021-04-09
张生
分享资料个
This version (03 Jan 2021 21:50) was approved by Robin Getz.The Previously approved version (12 Dec 2017 12:33) is available.Diff
EVAL-CN0398-ARDZ Shield Overview
CN0398 is a single supply, low power, high precision complete solution for soil moisture and pH measurements, including temperature compensation. The circuit is optimized for use with capacitive soil moisture sensors that are insensitive to water salinity and do not corrode over time. The circuit also measures soil pH and that function is suitable in a variety of applications.
The system is divided into three independent measurement front ends: pH, soil moisture, and temperature. After signal conditioning, the three channels share an AD7124-8, 24-bit sigma-delta (Σ-Δ) ADC. The AD7124-8, is a low power, low noise, completely integrated analog front end for high precision measurement applications. The device contains a low noise, 24-bit Σ-Δ analog-to-digital converter (ADC), and can be configured to have 8 differential inputs or 15 single-ended or pseudo-differential inputs. The on-chip gain stage ensures that signals of small amplitude can be interfaced directly to the ADC. The AD7124-8 uses the ADR3433,a low cost, low power, high precision CMOS voltage reference, featuring ±0.1% initial accuracy, low operating current, and low output noise. The ADR3433 provides both the 3.3 V reference and also the AVDD supply to the AD7124-8. The ADP7118-2.5 is a CMOS, low dropout (LDO) linear regulator which provides the power supply for the moisture sensor. The input to the ADP7118-2.5 is selectable via the jumper P10 as either 5 V or 7 V to 12 V coming from the Arduino-compatible platform board. The output of the ADP7118-2.5 is also selectable as 3.3 V or 5 V using the jumper on P8. The circuit uses the ADA4661-2, precision op amp to buffer the high impedance pH probe output and to drive the ADC. The ADA4661-2 is a dual, precision, rail-to-rail input/output amplifier optimized for low power, high bandwidth, and wide operating supply voltage range.
Equipment
  • EVAL-CN0398-ARDZ Evaluation Board
  • EVAL-ADICUP360 Base Board
  • PC with a USB port and Windows 7 (32-bit) or higher
  • Serial Terminal Software (Putty/TeraTerm or similar)
  • 7 V to 12 V/1 A dc power supply, or equivalent 7 V to 12 V/1 A bench supply
  • 3-wire moisture sensor
  • pH sensor
  • PT100 RTD
General Setup
  • The EVAL-CN0398-ARDZ (CN0398 Shield board) connects to EVAL-ADICUP360 base board
  • Power the system through the EVAL-ADICUP360 using external wall wart (7V-12V)
  • Connect the system to the PC using USB cable
  • Connect the sensors to the dedicated connectors
Test Setup Functional Block Diagram
Connectors and Jumper Configurations
Chip Select
The chip select is configurable to be routed to 3 general purpose I/O pins of the board, digital pins 8, 9 and 10, allowing multiple boards using SPI communications protocol to be stacked up.
As default, shunt is placed across pins 1 and 2 of connector P5 of EVAL-CN0398-ARDZ shield board which associates to the used of GPIO pin 10 of EVAL-ADICUP360.
Moisture Sensor Connector (P2)
Any 3-wire moisture sensor with a supply voltage of 3.3V or 5V, with an output voltage of 0-3V can be used with the EVAL-CN0398-ARDZ. The Vegetronix VH400 and Decagon Devices EC-5 moisture sensor was used during testing.
Moisture Sensor Power Supply Jumper Selection
If the moisture sensor requires a 5V power supply place the jumper as seen on the image below
VH400 moisture sensor was evaluated in this configuration
If the moisture sensor requires a 3.3V power supply, place the jumper as seen on the image below
EC5 moisture sensor was evaluated in this configuration. The two jumper selection P8 and P10 should always change together
RTD Connector (P1)
Possible RTD Configuration
3-wire PT100 RTD was used during evaluation, Excitation currents are set to 500uA. For the 4-wire configuration, Do not connect the RSENSE(+) wire.
pH Probe Connector (J1)
The EVAL-CN0398-ARDZ shield board comes with a BNC connector J1 for pH sensors as shown on the image below
The output of the pH sensor is bipolar and gives a maximum signal of ±414 mV at 25°C. The AD7124-8 operates from a single power supply therefore the pH probe should be biased above ground so that it is within the acceptable common-mode range of the AD7124-8. One of the integrated features of the AD7124-8 is its internal bias voltage generator that sets the common mode voltage of a channel to AVDD/2, or 1.65 V. This bias voltage from the ADC is applied to the pH probe shield and sets the output of the sensor 1.65 V ±414 mV at 25°C. For information about pH calibration, visit CN0398 Circuit Note.
Atlas Scientific pH probe was used during evaluation
Schematic, PCB Layout, Bill of Materials
EVAL-CN0398-ARDZ Design & Integration Files
  • Schematics
  • PCB Layout
  • Bill of Materials
  • Allegro Project
Software